AMETEK 9900RM Gas Analyzer for CEMS

    The new Model 9900RM is a single or multi-component gas analyzer that can be used alone or as an integrated part of a CEM (continuous emissions monitoring) system. The analyzer is 19” rack-mountable and supports Modbus/RS485 and Ethernet connectivity.

    The Model 9900RM can be configured to measure most gas species that absorb in the UV. Species typically measured directly by the Model 9900RM include: SO2, NO, NO2, H2S, COS, CS2, NH3, and BTX. Up to five components can be measured simultaneously making the system very economical for multicomponent analysis needs.

    The simple and robust design of the Model 9900RM is complemented by powerful data processing capabilities. The user-friendly keyboard enables programming of such variables as timing and frequency of local zero and span checks. Both analog and digital outputs are available and serial communications via Modbus protocol. An optional paramagnetic oxygen sensor can be included to provide O2 corrected concentrations.

When your requirement is for a rugged, single or multi-gas analyzer, free of interferences from water and CO2, the Model 9900RM is your answer.


Specifications
Species MeasuredSingle Species
Minimum Full ScaleMulti-Species
Minimum Full Scale
SO210 ppm20 ppm
H2S50 ppm100 ppm
NO50 ppm200 ppm*
NO2100 ppm200 ppm*
NOxn/a200 ppm*
* Lower ranges available upon request

Maximum full scale on all systems:100%
Methodology:Multiple wavelength, high resolution, non-dispersive ultraviolet
Optional O2:Integral paramagnetic sensor
Accuracy:Better than 1.0% of standard full scale range O2 : ±0.1%
Repeatability:±0.5% full-scale of standard ranges O2 : ±0.2% (reading varies by 0.05% due to A/D resolution)
Linearity:SO2: < ±1.0% full scale
NO, NO2, NOx, H2S: < ±1.5% full scale
O2: ±0.1%
Response Time:90% in 30 seconds or less
Number of Gases:1 (single species version)
Up to 5 (multi-species version)
Typical Sample Flow:0.25 - 5 l/min (0.6 - 10.5 SCFH)
Sample Gas Requirements:Noncorrosive, non-condensing, free from oil and with less than three micron particulates (0-95%, non-condensing)
Sample System Limits:5 psig max (without the dynamic pressure compensation option, the reported concentrations will be directly proportional to barometric pressure)
Pressure and Temperature Compensation:Optional
Ambient Conditions
Storage:-20 to 60C (-4 to 140F)
Operating:-20 to 50C (-4 to 122F) without Paramagnetic O2 sensor, 0 to 50C (32 to 122F) with Paramagnetic O2 sensor 0.95 Relative Humidity
Zero Drift:Typically better than 2% full scale standard ranges in 24 hours
Utility Requirements
Electrical:120 VAC (105 to 132 VAC), 47 to 63 Hz
240 VAC (209 to 264 VAC), 47 to 63 Hz
Power Consumption:< 200 W maximum start-up (from a cold start) with average power dependant on ambient temperature
Outputs:Four isolated Analog Outputs, each configurable to 0-20 mA, 4-20 mA, 0-5V DC, 1-5V DC
One configurable relay contact, Form A (SPST), normally open. Switching: maximum 240 V DC, 0.5 A DC, limited to 10 W. Carry: maximum 1.2 A DC
Inputs:Two isolated analog inputs, each configurable to 0-5 Volts or 0-20 mA.
Communications:Modbus RTU RS485 and Modbus TCP
Fast Ethernet - network connection or web-based server
Physical Dimensions (W x H x D):48.2 cm x 17.7 cm x 60.256 cm (19 in. x 7 in. x 23.7 in.)
Weight:18.2 kg (40 lb)